Calibration Procedure for Spectrophotometric Analysis of Plant Extract
- Begin by setting the spectrophotometer to a wavelength of 640 nm and select the % transmittance mode by pressing the A/T/C button, which will display the %T on the digital readout.
- Label one 3 mL syringe as "ethanol" and the other as "leaf extract."
- To prepare a blank sample, dispense 3 mL of ethanol into a clean glass cuvette, ensuring that any fingerprints are removed using a Kimwipe.
- Place the cuvette containing the ethanol blank into the spectrophotometer's cell holder and close the door securely.
- Press the 0 ABS/100%T key to calibrate the instrument, setting the blank to 100% transmittance.
- Remove the ethanol blank cuvette and prepare to test the plant extract. Dispense 3 mL of the filtered plant extract into a clean glass cuvette, again wiping off any fingerprints.
- Insert the cuvette filled with the plant extract into the spectrophotometer cell holder and close the door.
- Monitor the % transmittance reading on the display; aim for a reading between 30% and 40%. If the reading exceeds 40%, no dilution is necessary and you may proceed to further analyses. Conversely, if the reading is below 30%, dilution will be required.
- To dilute the extract, use the ethanol syringe to add 0.5 mL of ethanol to the cuvette. Seal the cuvette with your finger and gently invert it to mix the contents thoroughly.
- Re-insert the cuvette into the spectrophotometer and check the % transmittance reading. Continue to add 0.5 mL increments of ethanol, mixing thoroughly each time, until the % transmittance exceeds 30%, ensuring that you maintain a valid range for accurate measurements.